Most reference light sources are incandescent, running on very well regulated moderate/low voltage so as to reduce the effects of filament wear, and to get them to output Illuminant A ( (x,y)=(0.44758, 0.40745) - around 2856 K). The problem is that inexpensive light meters will read VERY differently at Illuminant A, which is what they were calibrated with, than they will at 5400K or 6500K.
So even if you calibrated your setup with an accurate incandescent light source, you would not get accurate readings from an LED flashlight with any of the Extech or similar meters. You'd be better off with an LED reference, which shouldn't be too hard to build. Or just use a known LED flashlight with a regulated power supply...
To get accurate measurements at different frequencies you have to spend a LOT more than the price the meter you are using, the most accurate being a spectroradiometer (I own two).
William