    Default AW IMR Cell Testing

    Greetings Everyone

    I purchased a number of AW IMR cells a month or so ago for the purpose of doing similar tests but had not gotten around to it. Well, last Friday I found myself standing in front of my Maccor battery tester and inspired by DFiorentino's recent post showing his test results on a variety of IMR cells. So I grabbed my box of AW IMR cells and went to work. I know that this is somewhat of a duplication of DFiorentino's efforts, but I already had the cells, so I figured I might as well run the tests.

    Here are the cells that were tested:

    14500 (rated 600 mAh)
    16340 (rated 550 mAh)
    18350 (rated 700 mAh)
    18490 (rated 1100 mAh)
    18650 (rated 1600 mAh)
    26500 (rated 2300 mAh)

    Here are some of the test details.

    Test Setup

    All tests were performed on a Maccor Seriers 4300 battery tester. The Maccor has separate leads for voltage sensing. If connected to a cell properly, this eliminates voltage errors due to contact and lead resistance. All cells had two nickel tabs spot welded to each lead. One tab was connected to the voltage sense lead and the other to the current leads. In this way even the resistance of the spot welded nickel tab should not affect the results.

    Test Procedure

    All cells were subjected to one initial "break-in" cycle where they were charged at a C/2 rate to a C/20 cut-off, allowed to rest for 1 hour then discharged at C/2 to 2.5 V. This first break-in cycle is not reported in any of the curves below.

    Following the break-in cycle, all tests were performed with the following procedure:

    1.) charge at C/2 to 4.2 V with a current cut-off of C/20
    2.) rest for 1 hour
    3.) discharge at specified rate to 2.5 V
    4.) rest for one hour
    5.) repeat steps 1-4 with a different discharge current

    Limitations

    One cell is hardly what I would call a statistically significant sample. There is certainly going to be variability in cell performance, so please recognize this limitation. Unfortunately, I do not have the time to repeat this work on multiple samples. These results are what you might expect for new cells. Exactly how the cells degrade will depend on the cell design, manufacturing quality and your use/abuse pattern.

    In other words, your mileage may vary.

    Presentation of the data

    It is always difficult to figure out how to present a lot of data so that it is useful for a broad audience. In an attempt to cover my bases and provide something for everyone, I am presenting Ragone plots and discharge curves. I decided to present the discharge curves in two ways. The first series of plots represents one cell size per graph and shows discharge curves for the following conditions:

    0.5 A, 1 A, 2 A and 4 A (14500, 16340 and 18350)
    0.5 A, 1 A, 2 A, 4 A and 6 A (18490)
    0.5 A, 1 A, 2 A, 4 A, 6 A and 8 A (18650)
    0.5 A, 1 A, 2 A, 4 A, 6 A, 8 A, 10 A and 15 A (26500)

    The second series of plots is an attempt to compare the performance of the cells in a normalized way. These plots show the C/2, 1C, 2C, 3C, 4C and 5C discharge performance normalized to the rated capacity of the cell.

    Ragone Plots

    If you are unfamiliar with Ragone plots, see my Intro to Ragone Plots thread.

    Ragone plots are great ways to represent a lot of information in a compact form. If you have a regulated light, then Ragone plots are going to probably provide you with all of the information you need. I discuss how to use these Ragone plots in post #24 further down in this thread.

    The first plot below shows the actual power vs energy response for all of the cells. The second plot is normalized to cell volume.

    Please note that these plots are currently incomplete. I still need to test the 18650 and 16340 cells, and I need to run the 18490 and 26500 cells at higher discharge power to establish their power limits. I will update these plots as additional data is collected.





    Constant Current Discharge Results per Cell Size















    Constant C-Rate Discharge Results Normalized to Rated Capacity













    I find this last plot showing the 5C rate comparison very interesting. It is clear that the 16340 cell does not hold up as well as the others. You also see that performance of the 14500 and 18490 are very similar, as are the 18350 and the 26500. It might mean that these similarly performing cells are made by the same supplier, or it might mean nothing at all.

    I hope that you guys find this useful. If there are any additional discharge conditions you would like to see, let me know and I will do my best to add them to these plots.

    Cheers,
    BG
    Last edited by Battery Guy; 11-17-2011 at 05:36 PM. Reason: added info
    Friends don't let friends use alkaline batteries.

  2. #2
    Administrator Kestrel's Avatar
    Join Date
    Oct 2007
    Location
    Willamette Valley, OR
    Posts
    6,070

    Default Re: AW IMR Cell Testing

    Very nice, thank you.

    Quote Originally Posted by Battery Guy View Post
    I find this last plot showing the 5C rate comparison very interesting. It is clear that the 16340 cell does not hold up as well as the others. You also see that performance of the 14500 and 18490 are very similar, as are the 18350 and the 26500. It might mean that these similarly performing cells are made by the same supplier, or it might mean nothing at all.
    Any similarity/differences in appearance regarding the positive nipples?

  3. #3
    Flashaholic* Battery Guy's Avatar
    Join Date
    Apr 2010
    Location
    Portland, Oregon
    Posts
    807

    Default Re: AW IMR Cell Testing

    Quote Originally Posted by Kestrel View Post
    Very nice, thank you.


    Any similarity/differences in appearance regarding the positive nipples?
    I did not look close. When I get a chance I will take some photos and post them.
    Friends don't let friends use alkaline batteries.

  4. #4
    Flashaholic* kiely23+'s Avatar
    Join Date
    Mar 2006
    Location
    Germany
    Posts
    509

    Default Re: AW IMR Cell Testing

    very good!
    thank you...
    "haben ist besser als brauchen"

  5. #5

    Default Re: AW IMR Cell Testing

    thanks for this very informative post. I think the most interesting thing is the comparison between 16340 and 18350. There's been quite a few legos put together that is using a 16340 to power triple and quad xp-g dropins and it's clear from the graphs that pushing the cell at 4A is just too much. Of course some of these dropins are capable of going well over 4A.
    Who needs to see the light at the end of the tunnel when you have friends on CPF?
    My flashlight videos: http://www.youtube.com/playlist?list...9TIYcGeuBXa5m0

  6. #6
    Flashaholic* 45/70's Avatar
    Join Date
    Oct 2005
    Location
    Rural Ohio
    Posts
    2,800

    Default Re: AW IMR Cell Testing

    Nice job, BG.

    I've been using LiCo cells for quite a few years, but only recently have acquired four of AW's 18650 IMR cells. I have just never needed what these cells are best at, offering good performance at high current rates (1C+). It will be interesting to see how the IMR 18650 cells perform.

    In the meantime, I am now also in need of IMR 16340 and 14500 cells. Unfortunately for this particular application (Jetbeam RRT-0, which can utilize either size cell), they have to have at least a "button top", if not an actual "nipple". This kind of narrows the choices down quite a bit.

    I'm torn between using AW's "protected" LiCo cells, or using IMR cells. The LiCo cells work great, but I really don't like running them at 1.7A. On the other hand, in addition to this light incorporating the "cheap" solution to polarity protection (a recessed positive contact, instead of a diode), it also offers no over discharge protection, as most of my other lights do. These two "features" are unbecoming of a $100 pocket light, IMO.

    Anyway, thanks for performing all the work involved and providing the resultant information. I will keep an eye out for the 18650 update. From experience, I'm expecting they will also perform well. I'll also add, that considering the size of a 16340 cell, I'm not too surprised they falter a bit at high current.

    Dave

  7. #7
    Flashaholic* Battery Guy's Avatar
    Join Date
    Apr 2010
    Location
    Portland, Oregon
    Posts
    807

    Default Re: AW IMR Cell Testing

    Quote Originally Posted by 45/70 View Post
    In the meantime, I am now also in need of IMR 16340 and 14500 cells. Unfortunately for this particular application (Jetbeam RRT-0, which can utilize either size cell), they have to have at least a "button top", if not an actual "nipple". This kind of narrows the choices down quite a bit.
    With the exception of the 26500 cell, all of the AW IMR cells that I tested had a relatively small button top for the positive terminal. I remember this clearly because it was very challenging to spot weld two separate leads to positive terminal for voltage sensing.

    Lighthound now has AW IMR 18650 cells in stock. I ordered two yesterday for testing, and hope to have the results posted in a week or so.

    Also, when I purchased the IMR cells I also purchased samples of the rest of AW's line of LCO cells. I hope to perform similar tests on these and post the results in a new thread in a few weeks.

    Cheers,
    BG
    Friends don't let friends use alkaline batteries.

  8. #8
    Flashaholic* 45/70's Avatar
    Join Date
    Oct 2005
    Location
    Rural Ohio
    Posts
    2,800

    Default Re: AW IMR Cell Testing

    Quote Originally Posted by Battery Guy View Post
    With the exception of the 26500 cell, all of the AW IMR cells that I tested had a relatively small button top for the positive terminal. I remember this clearly because it was very challenging to spot weld two separate leads to positive terminal for voltage sensing.
    Yeah, as I said, the selection of 16340 or 14500 IMR cells with nipples, is pretty much limited to either AW's, or AW's. I believe there may be some "CrapiFire" type cells available, but more and more have pretty much given up on cells of this type. I've sure bought and used enough of them to determine their quality, ie. low.

    I did check into shao's cells in the MarketPlace, but unfortunately, the only cells that have nipples are a few of the 16340 cells, and this seems rather random, as most do not seem to have them. On top of that, the 16340 cells don't seem to perform very well. All of his other cells are flat tops, or minimal button tops.

    I always appreciate tests of new cells. There are many to research that help with making a decision when purchasing cells for specific applications. On the other hand, there are very few tests of used cells, some, but not many. It is my opinion that these tests better reveal the true ability and performance of any chemistry cell.

    I don't now whether you've seen this thread that I started a while back. It involves testing some older used AW cells and, in most cases, some used, but not as much, or as old "CrapiFire" cell types. The results are typical of what I've experienced in my nearly 7 years, using LiCo cells in flashlights/torches. As I state in the thread, considering both the number of cycles, and the age of the cells tested, one could easily come to the conclusion that the AW cells would last at least four times as long as the competitors presented.

    Dave

  9. #9
    Flashaholic
    Join Date
    Aug 2011
    Location
    MD
    Posts
    336

    Default Re: AW IMR Cell Testing

    In the first graph,
    note that discharge current, I, is not a linear function of measured capacity
    so to predict capacity C from I
    take the log of I to make the relationship linear


    Measured ....discharge ........log of I
    capacity, C ....current, I .........LI
    0.35 ............2 .................. 0.301029996
    0.25 ...........4 ...................0.602059991


    Have Excel calculate the slope and intercept
    -0.332192809 =slope
    0.45 =intercept


    So C should equal
    -0.332192809 times Log(I) +
    0.45


    confirm
    I LI ................C
    2 0.301029996 0.35
    4 0.602059991 0.25


    so for a discharge current, I, of
    3
    take the log
    0.477121255 =log I

    then the capacity should be C =
    -0.332192809 times
    0.477121255 plus
    0.45
    equals
    0.29150375

    and not
    0.3
    as you would expect from a linear relationship
    between C and I

    In this case the difference is only
    2.8 percent but this may be
    larger with other batteries
    Last edited by xul; 11-03-2011 at 10:33 AM.

  10. #10
    Flashaholic* Battery Guy's Avatar
    Join Date
    Apr 2010
    Location
    Portland, Oregon
    Posts
    807

    Default Re: AW IMR Cell Testing

    xul

    Sorry buddy, but you totally lost me. What exactly are you trying to point out in the first graph?

    Cheers,
    BG
    Friends don't let friends use alkaline batteries.

  11. #11
    Flashaholic
    Join Date
    Aug 2011
    Location
    MD
    Posts
    336

    Default Re: AW IMR Cell Testing

    Quote Originally Posted by Battery Guy View Post
    xul

    Sorry buddy, but you totally lost me. What exactly are you trying to point out in the first graph?

    Cheers,
    BG
    Sorry.
    I need to take a writing course.

  12. #12
    Flashaholic* Battery Guy's Avatar
    Join Date
    Apr 2010
    Location
    Portland, Oregon
    Posts
    807

    Default Re: AW IMR Cell Testing

    Quote Originally Posted by xul View Post
    Sorry.
    I need to take a writing course.
    I doubt that very much. I just did not understand what the calculations were for. Looks interesting. Please explain.

    Cheers,
    BG
    Friends don't let friends use alkaline batteries.

  13. #13
    Flashaholic
    Join Date
    Aug 2011
    Location
    MD
    Posts
    336

    Default Re: AW IMR Cell Testing

    Well, battery testing graphs may show current levels of 0.1, 1 and 10A. This much variation, two orders of magnitude, is better understood using a log scale rather than a linear scale.

    The trouble comes if you want to know performance for an intermediate value of current. If you linearly interpolate between values you will be significantly off, i.e., halfway between the 1A line and the 10A line is 3A, logarithmically speaking, not ~5A.

    My spreadsheet is for doing this tedious calculation, back and forth between linear and log scales.

    Nailing down battery performance can be really difficult, at least for the reason of the exponential relationships between almost everything.
    And we all think linear.

    Too bad the sliderule people are out of business. These things have a one-for-one correspondence between linear and log scales.
    Last edited by xul; 11-05-2011 at 09:53 AM.

  14. #14
    Flashaholic* Battery Guy's Avatar
    Join Date
    Apr 2010
    Location
    Portland, Oregon
    Posts
    807

    Default Re: AW IMR Cell Testing

    Quote Originally Posted by xul View Post
    Well, battery testing graphs may show current levels of 0.1, 1 and 10A. This much variation, two orders of magnitude, is better understood using a log scale rather than a linear scale.
    Ah, I understand now. I was thinking about testing these under constant power conditions and plotting Ragone plots, which are log Power versus log Energy. Maybe I should do that.
    Friends don't let friends use alkaline batteries.

  15. #15
    Flashaholic
    Join Date
    Aug 2011
    Location
    MD
    Posts
    336

    Default Re: AW IMR Cell Testing

    Quote Originally Posted by Battery Guy View Post
    Ah, I understand now. I was thinking about testing these under constant power conditions and plotting Ragone plots, which are log Power versus log Energy. Maybe I should do that.
    With switch-mode regulators constant power is an option and it's generally different than constant resistance or constant current.

    Decide what the goal of your testing is and then design your test around that goal. If constant power regulators are commonly used then test results that relate to this mode will be useful for many people.

  16. #16

    Default Re: AW IMR Cell Testing

    Quote Originally Posted by xul View Post
    Well, battery testing graphs may show current levels of 0.1, 1 and 10A. This much variation, two orders of magnitude, is better understood using a log scale rather than a linear scale.

    The trouble comes if you want to know performance for an intermediate value of current. If you linearly interpolate between values you will be significantly off, i.e., halfway between the 1A line and the 10A line is 3A, logarithmically speaking, not ~5A.

    My spreadsheet is for doing this tedious calculation, back and forth between linear and log scales.

    Nailing down battery performance can be really difficult, at least for the reason of the exponential relationships between almost everything.
    Assuming an exponential relationship, such as Arrhenius, then it seems you ought to take natural logs, not base 10 logs. Using base 10 vice base e results in a constant shift error. Also, if different mechanisms operate at low vs high discharge current, then you will also have different activation energies, which can cause further problems with the interpolation, since it really becomes an extrapolation. I don't know if this is actually the case, but unless you have knowledge of operative mechanisms behind the exponential relationships, it is a risk. If I were to assume the same mechanism across the board, then I might use something like Lagrange interpolation or divided differences, and take advantage of the other data points available to me, not just the two adjacent ones.

    Your check that the slope and intercept values are correct also seems unusual. Of course the back check will give you exact results. You conducted a linear regression using two data points. Well, any two points define a straight line so of course you'll get back the same x,y points.

  17. #17
    Flashaholic
    Join Date
    Aug 2011
    Location
    MD
    Posts
    336

    Default Re: AW IMR Cell Testing

    If I had the software I'd do more sophisticated analyses.
    I tried to download a free program, I think it was 'R', but something went wrong. Computers are great when they actually work.

  18. #18
    Flashaholic* Battery Guy's Avatar
    Join Date
    Apr 2010
    Location
    Portland, Oregon
    Posts
    807

    Default Re: AW IMR Cell Testing

    I just started the tests on the AW IMR 18650 cells today, and hope to post results by Friday. I also started constant power discharge tests on the other cells, so I hope to post a partial Ragone plot on Friday as well.

    Cheers,
    BG
    Friends don't let friends use alkaline batteries.

  19. #19
    Flashaholic
    Join Date
    Aug 2011
    Location
    MD
    Posts
    336

    Default Re: AW IMR Cell Testing

    Check the variation between several identical cells.
    If you see a +/- 5% variation between cells and a +/- 5% variation between brands, the cell variation may be masking the brand variation.

  20. #20
    Flashaholic* Battery Guy's Avatar
    Join Date
    Apr 2010
    Location
    Portland, Oregon
    Posts
    807

    Default Re: AW IMR Cell Testing

    Quote Originally Posted by xul View Post
    Check the variation between several identical cells.
    If you see a +/- 5% variation between cells and a +/- 5% variation between brands, the cell variation may be masking the brand variation.
    As stated in the limitation section in the original post, I acknowledge that a sample of one is far from being statistically significant. Unfortunately, I do not have the time or resources to do testing on multiple cells.

    Cheers,
    BG
    Friends don't let friends use alkaline batteries.

  21. #21
    Flashaholic* Battery Guy's Avatar
    Join Date
    Apr 2010
    Location
    Portland, Oregon
    Posts
    807

    Default Re: AW IMR Cell Testing

    Greetings Everyone

    I have added data for the AW IMR18650 cell to the first post.

    Cheers,
    BG
    Friends don't let friends use alkaline batteries.

  22. #22
    Flashaholic* jasonck08's Avatar
    Join Date
    Sep 2007
    Location
    Redding, CA
    Posts
    1,516

    Default Re: AW IMR Cell Testing

    You should also try some higher discharge currents, after all most of these cells are rated 8-10C.

  23. #23
    Flashaholic* Battery Guy's Avatar
    Join Date
    Apr 2010
    Location
    Portland, Oregon
    Posts
    807

    Default Re: AW IMR Cell Testing

    Quote Originally Posted by jasonck08 View Post
    You should also try some higher discharge currents, after all most of these cells are rated 8-10C.
    I am planning on it. I will be running constant current tests at higher discharge currents to add to the above tests, and I am presently doing constant power tests so that I can construct Ragone plots for these cells.

    It is a lot of testing and takes a lot of time, so be patient.

    Of course, I probably won't be running the 16340 any higher. At 5C it is already a pretty poor performer.

    Cheers,
    BG
    Last edited by Battery Guy; 11-11-2011 at 06:25 PM.
    Friends don't let friends use alkaline batteries.

  24. #24
    Flashaholic* Battery Guy's Avatar
    Join Date
    Apr 2010
    Location
    Portland, Oregon
    Posts
    807

    Default Re: AW IMR Cell Testing

    Greetings Everyone

    I have added Ragone plots to the first post in this thread. They are not complete yet, but I will update them as more data is collected.

    I thought that I would take a moment to talk about how to use a Ragone plot to select an appropriate cell, or determine how fast a particular cell should be discharged.

    Let's consider the Ragone plot below:



    At this point, only the curves for the 14500 and 18340 cells are complete, so we will only consider those cells for the purposes of this discussion.

    Every cell has a characteristic Ragone plot that has the following similar features:

    1.) a relatively vertical, straight zone at low power where the amount of energy available is relatively insensitive to discharge power
    2.) a relatively flat, horizontal zone at high power where the amount of energy is very sensitive to discharge power
    3.) a transition zone between these zones

    The important point to remember is that you only want to use the battery in the first section, i.e. the vertical, straight zone where the amount of energy available is relatively insensitive to discharge power. There are two reasons for this:

    1.) The difference between the total energy available and your actual discharge energy is dissipated as heat. The bigger the difference, the hotter the cell will get during discharge and the fewer cycles you will get out of it.
    2.) As cells age, the total capacity available decreases and the internal resistance increases. The result is that the Ragone plot shifts down and to the left. If you are operating a new cell in the transition zone or above, you will see a big decrease in performance as the cell ages.

    So let's take a look at the 14500 and 18340 cells in the Ragone plot above. For the 14500 cell, it looks to me like the first zone extends up to ~14 W, or just above the 6C discharge rate. So I would recommend that the 14500 cell be used for applications up to, but no more than about 12 W or 6C. For the 18340 cell, the first zone extends up to ~10 W, or just above the 3.3 C discharge rate. So I would recommend that this cell be used for applications up to, but no more than about 8 W or 3C.

    "But wait a minute, these cells are rated to 8C maximum continuous discharge!" I hear you cry. Well, you can see from the Ragone plot that both cells are indeed capable of 8C continuous discharge. However, you are going to take a pretty big hit on available discharge energy, you will be heating the cells up a lot, and you will likely notice pretty rapid performance degradation with continuous use at 8C, especially for the 18340 cell.

    Hope you find this useful. Let me know if you have any questions.

    Cheers,
    BG
    Friends don't let friends use alkaline batteries.

  25. #25
    Flashaholic* Battery Guy's Avatar
    Join Date
    Apr 2010
    Location
    Portland, Oregon
    Posts
    807

    Default Re: AW IMR Cell Testing

    I thought some of you might find it interesting to compare the AW IMR cells to a couple other commonly used cells, so I have added a conventional Eneloop AA and a Sanyo 2600 mAh 18650 lithium-ion (typical laptop cell) to the Ragone plot below:



    Note that the Sanyo 18650 cell has a relatively small transition zone. This is because the cell has a PTC that kicks on at discharge rates of ~3C.

    Cheers,
    BG
    Friends don't let friends use alkaline batteries.

  26. #26
    Flashaholic* 45/70's Avatar
    Join Date
    Oct 2005
    Location
    Rural Ohio
    Posts
    2,800

    Default Re: AW IMR Cell Testing

    Quote Originally Posted by Battery Guy View Post
    Note that the Sanyo 18650 cell has a relatively small transition zone. This is because the cell has a PTC that kicks on at discharge rates of ~3C.
    Heh, I was wandering what the "crook" in the Sanyo's plot was. I sometimes observe similar anomalies when my cat jumps up on the testing area, LOL!

    Looks like the AW IMR 18650 performed as expected. I'm preferring these to my Samsung 30A cells, even though they don't run as long overall.

    With download's Pocket Rocket, the IMR cells will run down pretty much until discharged (3.60-3.70 Volts OC) on "high" (~3A), before the voltage warning kicks in, which is at ~3 Volts. The Samsung's start to fail at around 3.80 Volts OC, so the runtime actually works out to being close to the same, on high. While at lower levels (150 and 600mA) the Samsung's definitely offer superior runtime, I can't help but think that the 30A cells are not going to last as long with frequent 3A loading, as the IMR cells will.

    Thanks for adding the Ragone plots, BG. I find these very interesting, if not useful, as well. I have an offline collection of all of your Ragone plots for reference.

    Dave

  27. #27
    Flashaholic* Battery Guy's Avatar
    Join Date
    Apr 2010
    Location
    Portland, Oregon
    Posts
    807

    Default Re: AW IMR Cell Testing

    Quote Originally Posted by 45/70 View Post
    Thanks for adding the Ragone plots, BG. I find these very interesting, if not useful, as well. I have an offline collection of all of your Ragone plots for reference.
    I am really glad somebody finds them useful because they are a bit of a pain to put together! There is A LOT of test data summarized in one Ragone plot.

    I think Ragone plots are absolutely fantastic tools. For a single cell, you can see very clearly what the performance limits are and where you should be operating. They are also great for comparing different cells.

    One shortcoming of Ragone plots is that they don't tell you operating voltage of a cell during discharge, and this is important for a lot of hotwire and similar unregulated mods. Therefore, the humble constant current discharge curves still have their place.

    Cheers,
    BG
    Friends don't let friends use alkaline batteries.

  28. #28
    Flashaholic* 45/70's Avatar
    Join Date
    Oct 2005
    Location
    Rural Ohio
    Posts
    2,800

    Default Re: AW IMR Cell Testing

    Quote Originally Posted by Battery Guy View Post
    I am really glad somebody finds them useful because they are a bit of a pain to put together! There is A LOT of test data summarized in one Ragone plot.
    Absolutely, and I understand that there is a lot of work that goes into making them!

    One shortcoming of Ragone plots is that they don't tell you operating voltage of a cell during discharge, and this is important for a lot of hotwire and similar unregulated mods. Therefore, the humble constant current discharge curves still have their place.
    Yeah, well, I have a lot of the simple constant current discharge plots from many sources, as well. Fortunately, I am able to custom make my own, for the cells that I have, with my CBA.

    Dave

  29. #29
    Flashaholic* Battery Guy's Avatar
    Join Date
    Apr 2010
    Location
    Portland, Oregon
    Posts
    807

    Default Re: AW IMR Cell Testing

    Quote Originally Posted by 45/70 View Post
    Yeah, well, I have a lot of the simple constant current discharge plots from many sources, as well. Fortunately, I am able to custom make my own, for the cells that I have, with my CBA.
    I was looking into getting a CBA, but it seems that a major shortcoming is that it does not have separate voltage sense leads. This was a killer issue for me, especially for high current testing. When I contacted the manufacturer, they told me that their software takes care of voltage errors caused by connection resistance, but I remain skeptical. Any thoughts on that?
    Friends don't let friends use alkaline batteries.

  30. #30
    HKJ's Avatar
    Join Date
    Mar 2008
    Location
    Copenhagen, Denmark
    Posts
    9,187

    Default Re: AW IMR Cell Testing

    Quote Originally Posted by Battery Guy View Post
    I was looking into getting a CBA, but it seems that a major shortcoming is that it does not have separate voltage sense leads. This was a killer issue for me, especially for high current testing. When I contacted the manufacturer, they told me that their software takes care of voltage errors caused by connection resistance, but I remain skeptical. Any thoughts on that?
    In the pro version of the software you can add a resistive compensation, but it does not solve problems with varying contact resistance in the connector.
    Another issue it the voltage resolution, it is 10 mV. I believe this is a bit coarse when making graphs.
    My website with battery, charger, usb reviews, comparisons & information: https://lygte-info.dk/