Need creative ideas for measuring LED intensity

blasterman

Flashlight Enthusiast
Joined
Jul 17, 2008
Messages
1,802
Since I don't have a spectrometer I've been trying to come up with a way to measure relative intensity levels between specific LEDs that has little bias in regards to spectrum as possible, and drawing a blank. Dealing with a lot of large reef tank builds and the sheer volume of offbrand LEDs I need some way to measure the relative intensity of LEDs. I can keep it as simple as using a Cree at a specific current, and then determining what intensity level 'X' LED is when using the same config given that 99% of everything else is going to be less efficient anyways. I really don't care about color readings - I want to determine how much less energy 'X' LED is emitting compared to the benchmark. That has to make things easier.

I thought about using a trick from the laser geeks and using a small solar panel, and using a multimeter to take voltage / current readings off the panel with bare LED's illuminating them at close range.

Problem with this is I need to take readings from both royals and various whites. Depending on composition solar cells start to get terribly non-linear around 460nm. When I start throwing in that math, along with actually trying to determine the offset I simply have too many varibles to keep things straight.

Any ideas appreciated.
 

idleprocess

Flashaholic
Joined
Feb 29, 2004
Messages
7,197
Location
decamped
I'm not an engineer, nor do I play one on youtube docu-dramas ... but if you had some color filters and a light meter, perhaps you could get some approximate readings on intensity at whatever wavelengths the filters work at then do some fuzzy sum-integration of said plots?
 

saabluster

Flashlight Enthusiast
Joined
Oct 31, 2006
Messages
3,736
Location
Garland Tx
Since I don't have a spectrometer I've been trying to come up with a way to measure relative intensity levels between specific LEDs that has little bias in regards to spectrum as possible, and drawing a blank. Dealing with a lot of large reef tank builds and the sheer volume of offbrand LEDs I need some way to measure the relative intensity of LEDs. I can keep it as simple as using a Cree at a specific current, and then determining what intensity level 'X' LED is when using the same config given that 99% of everything else is going to be less efficient anyways. I really don't care about color readings - I want to determine how much less energy 'X' LED is emitting compared to the benchmark. That has to make things easier.

I thought about using a trick from the laser geeks and using a small solar panel, and using a multimeter to take voltage / current readings off the panel with bare LED's illuminating them at close range.

Problem with this is I need to take readings from both royals and various whites. Depending on composition solar cells start to get terribly non-linear around 460nm. When I start throwing in that math, along with actually trying to determine the offset I simply have too many varibles to keep things straight.

Any ideas appreciated.

Maybe take one of the other tricks from laser geeks and use a TEC. Coat it with flat black BBQ grill paint and mount it to a heatsink. Then measure the voltage off that.
 

slebans

Enlightened
Joined
Mar 1, 2010
Messages
457
Location
Moncton, NB Canada
I remember seeing apps/add-ons for the IPhone and other smart phones that would be suitable for your application.

Stephen Lebans
 

MikeAusC

Enlightened
Joined
Jul 8, 2010
Messages
995
Location
Sydney, Australia
Since I don't have a spectrometer I've been trying to come up with a way to measure relative intensity levels between specific LEDs that has little bias in regards to spectrum as possible, and drawing a blank. Dealing with a lot of large reef tank builds and the sheer volume of offbrand LEDs I need some way to measure the relative intensity of LEDs. . . . . . . I really don't care about color readings - I want to determine how much less energy 'X' LED is emitting compared to the benchmark. . . . . .

So you want Radiometric measurements that measure the total radiant energy, without weighting according to the eyes' sensitivity vs colour ?
 

Kinnza

Enlightened
Joined
Nov 15, 2005
Messages
223
Location
Spain
Before having the spectroradiometer, I used a cheap spectroscope (you can use a difraction grating directly aswell). You take a pic of the spectrum and process it with sofware developed for this task, available for free on the net. It basically consist on process the pic (TIFF or RAW formats, so camera corrections and white balance can be discounted) to black/white image, so absolute scale can be inferred from gray scale. Then this info is plotted as a SPD graph.

The main difficulty with this method is to calibrate the wavelength of the pic. You can overcome it measuring with a back lighting of known spectrum which provides the calibration info for knowing the peaks, but which is discounted from the last process stage or well used only to determine the peaks wavelength and later process a pic of the SPD of interest alone, but now having the info about wl peaks).

Check this page, it has all the info required to do it, along with the free sofware or links to it. Very in deep analysis of how to get SPDs from pics.

I dont know of any built in sofware for Ipads or so that does it as Stephen says, but since I got the spectrometer I didnt search for it anymore. I wouldnt surprised if somebody has came with such solution. Would worth to search deep about it.
 

Kinnza

Enlightened
Joined
Nov 15, 2005
Messages
223
Location
Spain
Oh, forget to add how to know absolute radiance once you know the SPD. You need to have an integrating sphere or a goniometer, measure total flux and correct it for the LER of the SPD. For royal blues, it may become inaccurate when using standard light meters, which often are not well calibrated for wl below 460nm.

But if you only want comparative results, as the response is the same with the same sensor, you can get good enough data easily.
 

Kinnza

Enlightened
Joined
Nov 15, 2005
Messages
223
Location
Spain
I read your first post again, blasterman, and I noticed you are not interested on the total radiometric flux, but on intensity (mW/sr). In order to compare such magnitude, you just need to measure at a given distance of the optical axis with a light meter (for royal blue, better a good one) and correct it by its LER (obtained by its SPD).

Likely some software for digital camera or add-on for phone with camera could do it, but the latter probably would be limited by the sensibility on the blue range for short wl.
 
Last edited:

David_Campen

Enlightened
Joined
Jun 29, 2004
Messages
674
Location
California
Take a metal cylinder that is closed at both ends.
Paint the inside flat black.
Surround the cylinder with good thermal insulation.
Place a temperature measuring device underneath the insulation, mounted so that it is thermally well connected to the cylinder.
Make a hole in the cylinder so that the LED can be placed there emitting all of its light into the cylinder but with the LED thermally isolated from the cylinder.

The rate of change of the temperature of the cylinder will be proportional to the intensity of the light being directed into the cylinder. Compare an unkown LED against your reference LED and the ratio of the rate of change of cylinder temperature gives the ratio of the light intensities.
 

blasterman

Flashlight Enthusiast
Joined
Jul 17, 2008
Messages
1,802
you just need to measure at a given distance of the optical axis with a light meter (for royal blue, better a good one) and correct it by its LER (obtained by its SPD).

That's problem - how do I measure it? While there are many interesting suggestions here, most rely on CCD type photon counters, and these simply won't work because color CCDs use color cutoffs on their sensors. Digital camera sensors for instance typically have a 455nm blue filter, which means a light source at 455nm is going to read more accurate than one at 450nm or 460nm. I've already run into this problem when using 10watt 440nm LEDs given they don't record properly as standard ~455nm royals with a dSLR because the 455nm cutoff filter can't distinguish color/intensity below 455nm. So, any device using color CCD / CMOS is out. Not sure what the spectral weighting is for monochrome CCDs, at least one affordable for this project.

If I had a small solar cell that was fairly linear between 420-530nm I'd be happy, and that would work, but all the charts I've seen show they aren't. So, I need to start with a photon detector that doesn't care about color in the first place, at least in blue-green, and work from there.

The rate of change of the temperature of the cylinder will be proportional to the intensity of the light being directed into the cylinder

Very innovative, but I would think that temperature changes would be slight, and or take a very long time (kinetic transfer of energy, correct?). Plus, the assembly would have to be in a vacuum otheriwse ambient temperature influences would screw it up. (??)
 

MikeAusC

Enlightened
Joined
Jul 8, 2010
Messages
995
Location
Sydney, Australia
The "Clear" channel of the TCS3200 Colour Light to Frequency Converter has a response at 350nm that's 28% of its peak at 720nm. The sensitivity rises roughly linearly from 350 to 720nm.

You also have R, G and B channels available so you can make an assessment of the spectral distribution. The colour channel to be measured is set by applying DC to two input pins. The sensor area is 1mm x 1mm,

It has internal scalers so you can set the maximum to to 600, 120 or 12kHz. The Dark output is <2Hz. Many Multimeters have a frequency measurement capability.

It draws 2mA at 5 volts.

And all of this costs . . . . . . $6 !!!!!!!
 
Last edited:

MikeAusC

Enlightened
Joined
Jul 8, 2010
Messages
995
Location
Sydney, Australia
The TEMD5510 is just a Photodiode that has a response from 400 to 700nm, but between 400 and 520nm its repsonse is higher than the eye's.

The VTB8440 is just a photodiode that has a response from 330 to 720nm, but the spectral response is not published.
 

David_Campen

Enlightened
Joined
Jun 29, 2004
Messages
674
Location
California
The rate of change of the temperature of the cylinder will be proportional to the intensity of the light being directed into the cylinder

Very innovative, but I would think that temperature changes would be slight, and or take a very long time (kinetic transfer of energy, correct?). Plus, the assembly would have to be in a vacuum otheriwse ambient temperature influences would screw it up. (??)

Rate of change of temperature would not be slight if the cylinder is matched to the power of the LED.

Say the cylinder was 1" diameter X 1" long and constructed of 6.75 mil copper foil. The cylinder (with end caps) would weigh 4.6 grams. Copper has a heat capacity of 0.38 watt-second/gram-degC so if you put 2 watts into this cylinder the rate of change of temperature would be 1.1 degC/second.

Phenolic foam insulation has a thermal conductivity of 0.5 milliwatt/inch-degC. So if you insulated the cylinder with 1 inch thick phenolic foam the rate of heat loss would be 2.3 milliwatt at a 1 degC temperature differential.
 
Last edited:

slebans

Enlightened
Joined
Mar 1, 2010
Messages
457
Location
Moncton, NB Canada
I read your first post again, blasterman, and I noticed you are not interested on the total radiometric flux, but on intensity (mW/sr). In order to compare such magnitude, you just need to measure at a given distance of the optical axis with a light meter (for royal blue, better a good one) and correct it by its LER (obtained by its SPD).

Likely some software for digital camera or add-on for phone with camera could do it, but the latter probably would be limited by the sensibility on the blue range for short wl.

Kinnza, I just found the lowest price spectrometer I have ever seen. Generally, I have found pricing between $1499 and $2499. This one is just $749 and it originates from Canada - of all places. If you get a minute can you have a look and see what you think.
http://www.aseq-instruments.com/index.html

Stephen Lebans
 

MikeAusC

Enlightened
Joined
Jul 8, 2010
Messages
995
Location
Sydney, Australia
Kinnza, I just found the lowest price spectrometer I have ever seen. Generally, I have found pricing between $1499 and $2499. This one is just $749 and it originates from Canada - of all places. If you get a minute can you have a look and see what you think.
http://www.aseq-instruments.com/index.html

Stephen Lebans

Interesting - Toshiba don't specify the spectral response of the TCD1304 sensor below 400nm at all, although ASEQ can apply a custom calibration curve for each sensor to give reasonable accuracy.
 

slebans

Enlightened
Joined
Mar 1, 2010
Messages
457
Location
Moncton, NB Canada
Interesting - Toshiba don't specify the spectral response of the TCD1304 sensor below 400nm at all, although ASEQ can apply a custom calibration curve for each sensor to give reasonable accuracy.

This sensor is also used in the Ocean Optics' newer HR4000 models.

Stephen Lebans
 

SemiMan

Banned
Joined
Jan 13, 2005
Messages
3,899
Very cool spectrometers.

They even have low cost cosine corrected adapters. Would be pretty easy to whip up a basic goniophotometer.

Semiman
 

SemiMan

Banned
Joined
Jan 13, 2005
Messages
3,899
If I am interpreting what you are trying to do, then all you are looking at measuring is radiometric output of the LED?

1) You need some way of diffusing the light source to even out the distribution. You can make a poor mans integrating sphere with a box or sphere and some really good pure white flat paint. Even better, paint for the inside of signs ... >95% reflective and highly diffuse.

2) You need a somewhat linear light sensor. A wide bandwidth photo-diode. Do not operate it forward biased as a current source, but back-biased as a light based current modulator. It will be more linear in this mode. You may even be able to do it with a monocrystalline solar cell (just need a piece of one).