jayflash said:
Only cheaper...of course
I'm casting about for opinions & facts regarding a good value in a small DMM. Important features for this meter are (back) lighted display, 10A (20 peak) current range, larger readout desired, anything else is a +.
If something else comes close to a Fluke 112, is well made and, thereby, safe on <600v, for less $$, I'd appreciate your input. Otherwise, I may just get another 112.
You're paying a great deal for safety features on Fluke. There are two fuses inside a Fluke and each of these fuses cost more than a cheap DMM. Both fuses (0.44A & 11A) are rated 1000v and 10,000A+ interruption current, which relates to the ampere capability of the power source.
One thing a 10A (1000A IR) and 10A (170,000A) fuses have in common is that if you accidentally short the ammeter across a car battery, they will both blow and cost you money for a replacement fuse.
But, let's say a tech accidentally try to measure the voltage across two phases on the secondary of a 5,000kVA 480v transformer while the meter leads are in ammeter ports. The proper IR and voltage rated fuse will blow and contain the force w/ in the fuse.
The 10A 1000A(IR) 250V fuse won't be able to stop the current, which could mean the meter explodes violentlycausing an arcover between busbars, which then creates a massive blast and continue to melt things down until the utility company's protection equipment on the 12kV side of the transformer shuts it off.
I'm talking about working on something like this:
So.. since current measurements I make are mainly low voltage DC and the short circuit current of the source is not that high, I dedicate a cheapie for current measurement.
Comparing the readings in voltage mode, the accuracy is decent. I like my Flukes (I have three), but I admit the cheapie is just fine for many uses.
That CEN-TECH cheapie was like $5.00 at Harbor Freight
http://www.harborfreight.com/cpi/ctaf/displayitem.taf?Itemnumber=90899
If you want backlight, they have another cheapie w/ backlight
http://www.harborfreight.com/cpi/ctaf/Displayitem.taf?itemnumber=92020
I recommend you buy two to prevent wear and tear on the rotary switch. Leave one permanently dedicated to "10A" and the other one to 2,000mV or 2V and juggle between the meters, not "modes".
Since you already have a Fluke 112, you can open the cheapie and calibrate it to 112 using the only potentiometer inside. Just apply a 1.5v battery in 2000mV range and adjust until the display matches the Fluke. It might not even be necessary if yours is close enough.