Nothing is wrong with your DMM. Different values of resistors are used for each setting, so getting a lower current reading on the lower current scale is possible especally on an older DMM. The newer DMM's are a lot closer, or exact in some cases. Depends on manufacturer and $$$ of the meter. The lower the cost, the more likely this type of thing is to happen. The measure the voltage drop across an internal resistor to calculate current. The 10A scale probably uses a resistor in the milliohm range, the 200 ma scale probably uses a resistor in the ohms range. At such a low current the difference in resistance of the 2 scales is significant. I would trust the 10A scale 110ma reading over the 200ma scale 67ma reading. You need to know what the limitations of the measuring/test equipment you are using. Use another DMM on the ohms scale and measure the ohms across the 10A and 200ma scales and you will see what I mean.